Hi Carl,
a MONO coil is divided into 4 sections. Each of the parts are individually critical damped. A second standard MONO coil is compared to the first one.
The coil is driven as usual on both ends (not center-tapped).
The following picture (1) shows the damping voltages with no target presented.
The second picture (2) shows the comparison of the target/no-target response.
It could be useful to find the reason for the anomaly coil current with the standard coil, which delays the decay curve. This is totally absent with the bbsailor's proposal. I am not sure, whether this effect can be measured and verified in a real practical configuration.
Aziz
Announcement
Collapse
No announcement yet.
New PI technology proposals.
Collapse
X
-
340 ohms for the 300uH mono, and 235 ohms each for the CT coil.Originally posted by bbsailor View PostSince you were able to obtain a 1 us improvement in discharge time using two damping resistors, I suspect that the effective value of these two resistors in series across 10 turns each was a little higher than a single damping resistor across 20 turns. What were the values of the single Rd and the dual Rds?
I think you'll have a rapidly diminishing improvement that isn't worth the extreme pain of building such a coil!What if you damped the coils at 5 turns each effectivly having the coil taps connected to 4 series damping resistors? I suspect that the combined value of this 4-resistor set might be a little higher than with 2 resistors in series thus further improving the discharge time.
I guess if you extended this to it's logical conclusion, you could have a damping resistor on each turn. That might really damp the flyback?
I think all this can be determined with an IB coil pair, and maybe even with an ordinary mono coil.Then we could know the true role of the flyback pulse in (1) stimulating the target, (2) being a necessary artifact that must be quickly dispensed with or (3) if it produces some other measurable result related to detecting the target?
- Carl
Leave a comment:
-
Now that's clever.Originally posted by simonbaker View PostIf flyback can be damped faster with lots of resistors in coil, why not make a coil with flat spiral windings, use wire with no insulation, embedded in a fairly conductive layer of graphite? Infinite number of resistors!
Leave a comment:
-
Yes, that's what I'm doing.Originally posted by Aziz View PostThe center-tapped coil should be damped on each half with individual damping resistors instead of my proposal. This will give a faster decay.
I'm using PVC insulated wire. We have teflon as well, but I didn't have a spool handy. That's also why I wound a mono coil, for a 1-to-1 comparison.The coil wire should have an insulation to keep the interwire capacitance low. Then the capacitive coupling between the half coils will be low.
Well, that's true only if the ringing perfectly cancels. But it doesn't.As long as the ringing is totally in-phase, this shouldn't be a problem with differential signalling. But best would be to avoid any ringings.
Well, I looked but it must be in one of those hidden forums. Post the results here so everyone can see them.If you are interested on some amazing results, please visit the Oz-forum. I just posted there.
- Carl
Leave a comment:
-
Originally posted by Carl-NC View PostAziz,
I've built & tested the CT coil, plus an otherwise identical mono coil. Specs are:
Mono:- 20 turns of 24awg
- 290mm diameter
- 299uH / 4 ohms
- 10 + 10 turns of 24awg
- 290mm diameter
- 77uH / 1.1 ohms and 76uH / 1.1 ohms
These oscillations are common-mode in-phase so that when you look differentially, they mostly cancel but not completely. This is likely sensitive to coil matching & parasitic matching, which means that in reality it will be all over the place. So you end up with a faster decay but with oscillations.
I can, however, change the damping to that of a traditional two-coil PI front-end, where each coil is individually parallel-damped. In that case, I get rid of the oscillations and end up with a clean response both individually & differentially. It is about 1us faster that the mono response, which crosses 1v at a 5us delay.
So in this mode, we have the advantage of a faster di/dt due to the 77uH TX coil and we still get the full 300uH RX coil. Sounds good so far. But one thing I didn't mention before is that, yes, dB/dt field strength is proportional to di/dt, but it is also proportional to N, the number of coil turns. For the CT coil N is 1/2, so we better get at least a 2x di/dt to make up for this. I'm clipping my MOSFET so it's hard to say. Also, I don't have this hooked up to a full detector so I can't do target comparisons.
I would like to next wind up a CT coil using separate 10T windings so there is less interwinding coupling. Also need to rig up a differential receiver so I can run the whole metal detector.
- Carl
Carl
Since you were able to obtain a 1 us improvement in discharge time using two damping resistors, I suspect that the effective value of these two resistors in series across 10 turns each was a little higher than a single damping resistor across 20 turns. What were the values of the single Rd and the dual Rds?
What if you damped the coils at 5 turns each effectivly having the coil taps connected to 4 series damping resistors? I suspect that the combined value of this 4-resistor set might be a little higher than with 2 resistors in series thus further improving the discharge time.
I guess if you extended this to it's logical conclusion, you could have a damping resistor on each turn. That might really damp the flyback?
Then we could know the true role of the flyback pulse in (1) stimulating the target, (2) being a necessary artifact that must be quickly dispensed with or (3) if it produces some other measurable result related to detecting the target?
What do you think?
bbsailor
Leave a comment:
-
Hey Aziz:Originally posted by Aziz View PostHi Carl,
thanks for your practical efforts. The center-tapped coil should be damped on each half with individual damping resistors instead of my proposal. This will give a faster decay.
The coil wire should have an insulation to keep the interwire capacitance low. Then the capacitive coupling between the half coils will be low.
As long as the ringing is totally in-phase, this shouldn't be a problem with differential signalling. But best would be to avoid any ringings.
If you are interested on some amazing results, please visit the Oz-forum. I just posted there.
Regards,
Aziz
If flyback can be damped faster with lots of resistors in coil, why not make a coil with flat spiral windings, use wire with no insulation, embedded in a fairly conductive layer of graphite? Infinite number of resistors!
Maybe already done....
Cheers,
-SB
Leave a comment:
-
Hi Carl,
thanks for your practical efforts. The center-tapped coil should be damped on each half with individual damping resistors instead of my proposal. This will give a faster decay.
The coil wire should have an insulation to keep the interwire capacitance low. Then the capacitive coupling between the half coils will be low.
As long as the ringing is totally in-phase, this shouldn't be a problem with differential signalling. But best would be to avoid any ringings.
If you are interested on some amazing results, please visit the Oz-forum. I just posted there.
Regards,
Aziz
Leave a comment:
-
Only if you want it settled out to a steady-state level. At 3τ the coil current (and the B-field) will be 95% settled. But you can shut off the TX at any time... at 1τ the current will be 63% of the peak value. The drawback is that the coil current is still changing and can induce an undesired target reverse eddy current that subtracts from the (desired) turn-off eddies. But it's a negligible effect, and ramping the current up to several τ's just wastes power.Originally posted by 1843 View PostI think that establishing the magnetic field, requires a coil on-time at least 3 TC... Doesn't it?
- Carl
Leave a comment:
-
Are you doing this with an IB coil system? If yes, the the 100mv signal you see is likely due to TX-RX induction and not from any target.Originally posted by ZED View PostSB the thing i find intreaging is the different decay curves ,if we took a sample 10 u/s after the transmit starts we get a 100 milli volt signal and if we took a sample 10 u/s after the transmit stops we get a signal probably in the u/volts,the scope pic is certainly suggestive of the possibility of sampling during the on time for a better target responce.
Zed
- Carl
Leave a comment:
-
Aziz,
I've built & tested the CT coil, plus an otherwise identical mono coil. Specs are:
Mono:- 20 turns of 24awg
- 290mm diameter
- 299uH / 4 ohms
- 10 + 10 turns of 24awg
- 290mm diameter
- 77uH / 1.1 ohms and 76uH / 1.1 ohms
These oscillations are common-mode in-phase so that when you look differentially, they mostly cancel but not completely. This is likely sensitive to coil matching & parasitic matching, which means that in reality it will be all over the place. So you end up with a faster decay but with oscillations.
I can, however, change the damping to that of a traditional two-coil PI front-end, where each coil is individually parallel-damped. In that case, I get rid of the oscillations and end up with a clean response both individually & differentially. It is about 1us faster that the mono response, which crosses 1v at a 5us delay.
So in this mode, we have the advantage of a faster di/dt due to the 77uH TX coil and we still get the full 300uH RX coil. Sounds good so far. But one thing I didn't mention before is that, yes, dB/dt field strength is proportional to di/dt, but it is also proportional to N, the number of coil turns. For the CT coil N is 1/2, so we better get at least a 2x di/dt to make up for this. I'm clipping my MOSFET so it's hard to say. Also, I don't have this hooked up to a full detector so I can't do target comparisons.
I would like to next wind up a CT coil using separate 10T windings so there is less interwinding coupling. Also need to rig up a differential receiver so I can run the whole metal detector.
- Carl
Leave a comment:
-
During the on-time the dB/dt is less in compared to the transmit stop time, so the target response must be less, too.Originally posted by ZED View PostSB the thing i find intreaging is the different decay curves ,if we took a sample 10 u/s after the transmit starts we get a 100 milli volt signal and if we took a sample 10 u/s after the transmit stops we get a signal probably in the u/volts,the scope pic is certainly suggestive of the possibility of sampling during the on time for a better target responce.
Zed
I think there's something wrong with your experiment.
Regards,
1843
Leave a comment:
-
I think that establishing the magnetic field, requires a coil on-time at least 3 TC... Doesn't it?Originally posted by Qiaozhi View PostA current flows through a coil until the magnetic field is fully established, then the switch opens and the magnetic field collapses, creating a back-emf.
Regards,
1843
Leave a comment:
-
SB the thing i find intreaging is the different decay curves ,if we took a sample 10 u/s after the transmit starts we get a 100 milli volt signal and if we took a sample 10 u/s after the transmit stops we get a signal probably in the u/volts,the scope pic is certainly suggestive of the possibility of sampling during the on time for a better target responce.
Zed
Leave a comment:
-
Hello friends,
it seems, we could have some real breakthrough in the PI technology!!!
Provided that, the simulation results can be verified with the real measurements.

Aziz
Leave a comment:
-
It would be surprising to see different decay rates if both signals represented the "natural response" (stimulating signal gone, looking at natural decay) of the target. However, because everything is happening so fast, I believe we are looking at various "forced responses" (stimulating signal still present to some degree).Originally posted by ZED View PostMonolith my experiences with on time sampling have always shown that the target signal is always strongest at switch off,the pictured scope shot that i showed was from an enquire into how fast i could get the coil current to rise in a transmit coil.The reason for this was that i had an idea about sampling for a target signal at the start of the transmit while the ground signal was at its lowest amplitude,meaning "while the magnetic domains in the ground matrix were just starting to align themselves with the coil field".
If your getting a weaker signal at switch off then it sounds like the rate off coil field change is slower than when it was rising,how long does it take for the coil current in your setup to reach zero after switch off ?
Simonbaker
Yes the target signal after switch off was clipped,it went of the screen on the scope and the visual was uploaded to the pc and it showed it a little differently,cant remeber its peak value but it was in the volts range,also notice how the target signals have different decay rates from one another,interesting dont you think ?
Zed
It is tricky because we sort of have a natural response (the coil being turned on or off) of one system (the coil) being the forcing signal to another system (the target coin). Especially at current shutoff in the coil, we have two systems with very fast response, so the signal in the target coin is hard to interpret by looking at it.
Also, each time we go from magnetic field to current we take a derivitive, including on the way back to the receive coil -- I need to think about that a little. I'm now wondering if the target coin can make a nice signal as its current reverses direction from the up pulse to the down pulse -- in other words, do they tend to cancel each other, or could they enhance each other.
Need to think....
Cheers,
-SB
Leave a comment:
Leave a comment: