    Originally posted by metaldetector107 View Post
    dear max
    the pole is plastic pipe 80cm long and the rod
    or the thingy that i fixed the loop with pipe also os plastic
    i took them from an old PI metal detector called P-158
    only six nuts are metal which fixed the rod to the loop house
    the electronic circuit house is metal box


    so should i shelid the foil befor appling any tape on coil and is it should be very tight with no gaps??

    thanks
    Hi,
    the shield you can also not tape... but I usually do... what's important is that the shield foil will not move relative to coil windings, so stay all very strong and fixed... the best is , always, epoxing everything when the coil is made and shielded properly... that way the epoxy will block any further movement.

    The pole is plastic: OK check if strong enough... must not flex when you swing the coil...

    The nuts are few so I think all ok about that.

    The control box is metal so must stay at least at say 60-70cm from coil (more is better cause will influence less the coil in case of vibrations). If you cannot stay too far with it... use a plastic made box if possible. All metals are a problem in control box cause the box is big object and not too far from coil (it stays in the detection range... so must stay very fixed relative to coil or at vibrations you'll get a false signal).

    That's cause tgs/tgsl is very sensitive and if can detect a coke can at such distance easy (60/70cm) can detect also the movement of metallic control box. So stay far with it and very tight fixed or use plastic.

    Kind regards,
    Max

    Leave a comment:


  • metaldetector107
    replied
    TGSL

    dear max
    the pole is plastic pipe 80cm long and the rod
    or the thingy that i fixed the loop with pipe also os plastic
    i took them from an old PI metal detector called P-158
    only six nuts are metal which fixed the rod to the loop house
    the electronic circuit house is metal box


    so should i shelid the foil befor appling any tape on coil and is it should be very tight with no gaps??

    thanks

    Leave a comment:


  • Max
    replied
    Originally posted by metaldetector107 View Post
    Dear Max thanks soso much
    actually i am using pulse induction MD cable which has to seperate ground ...
    about green field i couldnot test it cause when i move the machine it sounds i went outside to make sure i am away from elelcrical sourses
    so it was same
    so my only problem now is the machine sound whenever i move it and
    wherever i am .
    one more thing before appling epoxy i measured the shiled ohms on of them was around 4ohms do you think is it the problem?

    thanks alot
    Hi,
    some suggestions on mechanical aspects...
    remember the coils are very sensitive to relative movements, so you must fix very well them : if used epoxy I think you fixed them ok.

    Then about the pole/rod you use for coil... must be rigid, very rigid thing: choose something very strong (wood is also ok if weight is not a concern).

    If metallic, like alluminium or iron or steel tube, be very careful cause ANY vibration of rod/pole relative to coils will create a false signal.

    Then I suggest, if use metallic pole, don't make it arrive to the coil proximity: I mean cut e.g. 20-25cm of last part of it near coil, then make a patch with a short wood rod or plastic tube... that will be much more relaxing cause when you'll sweep coil will be difficault having a false signal from pole torsion if last part is made of wood or plastic.

    The screw and wingnut you'll use at coil pivot you'll better choose in plastic made: nylon screws/plates/nuts are expecially nice there at pivot cause even if will move or vibrate (e.g. we you hit a rock) they will not give false signal to the head.

    Also, the box for electronics you'll better made of plastic; if made of metallic material you must keep at good distance from coil , this is essential even if pole/rod is very rigid.

    The cable of coils must be fixed to pole, lower part expecially cause movements could be detected as false signal too.

    All rigid, all fixed, all silent when you move. That's the way.

    These are just advices, then you could find good solutions for yourself with materials you have on hand or that could buy at your place.

    The shield resistance is not a big issue: few ohms is ok. The shield, of course, must be rigid so in epoxy like with coils... no movements allowed between shields and coils or will get false signal.

    Kind regards,
    Max

    Leave a comment:


  • metaldetector107
    replied
    tgsl

    Dear Max thanks soso much
    actually i am using pulse induction MD cable which has to seperate ground ...
    about green field i couldnot test it cause when i move the machine it sounds i went outside to make sure i am away from elelcrical sourses
    so it was same
    so my only problem now is the machine sound whenever i move it and
    wherever i am .
    one more thing before appling epoxy i measured the shiled ohms on of them was around 4ohms do you think is it the problem?

    thanks alot

    Leave a comment:


  • Max
    replied
    Originally posted by metaldetector107 View Post
    Dear Max
    i am happy to see u again i was waiting for you,,,
    actually it does reject nail ,pliser and other iron totaly. i dont use disc switch it is always on disc mode... the ferrite is the black thingy from radio antenna it is about 7cm long and 1cm diameter when adjusting GB i
    reject it till 2-3cm so i thing is good my problem today i took it to test it on soil
    it gave me more fale signal when lowerd to the ground and even when movig the machine, about shielding i have shelded both with the alaminium
    foil i took some out from coax cable forTV
    another thing max
    the TX freq is 14.7 and both coil 6mh tx 6.5mh rx
    but the frequncy from pin 1 of 4024 is around 14.5 which should be same as collector of 2n2907 as you metioned already
    so what do you think of these problem

    thanks very much
    Hi,
    the problem is probably related to shield only. The fact you can reject to 2-3cm a 7cm long ferrite rod (yes,it's that thing you find in am portable radios) x 1cm diameter (I think) means you have a good setting at pot and good phase at coil too.

    So, if problem is with e.g. hands or soil when you lower the coil and move around it's almost sure a shield related issue.

    Have you tested on green, possibly wet or humid grass ? It will go crazy if shield isn't the right way cause of static charges and capacitive effects between coils and soil or grass.

    Now, the shield must be made of e.g. aluminium foil... the kind used in tv cables (coaxial I think you mean, for vhf/uhf terrestrial tv signals) could sometimes be used but only if very thin one.

    Some, most of tv coaxial have heavy aluminium shield so that will somehow reduce your md performances (too heavy shield is not good cause will create too short conductive path near coils... means too eddy currents inside coil housing).

    But the problem of shield you have is related to connections of it I think: you must use two separate shields , one for tx coil, one for rx coil and two wires that then bounds just at cable or pcb. Do not e.g. make rx shield touch tx shield etc (create a short turn there) and always leave the gap in both shields to avoid short turn effects.

    Then you must check connections of shields to gnd of circuit, cause that's the purpose: make an electrical connection to gnd there to enstablish a fixed potential level otherwise your coils will see the influence of external charges.

    The shields will also provide masking of noises (rf noise, some electrical noise around etc) but this is another issue.

    Then check that connections, something is probably wrong there...fix it and it will be fine.

    If you notice, instead that connections are ok, hands will give no false signals your soil is probably heavy mineralized and you're reading falses from magnetic crystals in the matrix and such can be worse than the ferrite rod used in tests.

    Kind regards,
    Max

    Leave a comment:


  • metaldetector107
    replied
    tgsl

    Dear Max
    i am happy to see u again i was waiting for you,,,
    actually it does reject nail ,pliser and other iron totaly. i dont use disc switch it is always on disc mode... the ferrite is the black thingy from radio antenna it is about 7cm long and 1cm diameter when adjusting GB i
    reject it till 2-3cm so i thing is good my problem today i took it to test it on soil
    it gave me more fale signal when lowerd to the ground and even when movig the machine, about shielding i have shelded both with the alaminium
    foil i took some out from coax cable forTV
    another thing max
    the TX freq is 14.7 and both coil 6mh tx 6.5mh rx
    but the frequncy from pin 1 of 4024 is around 14.5 which should be same as collector of 2n2907 as you metioned already
    so what do you think of these problem

    thanks very much

    Leave a comment:


  • Max
    replied
    Originally posted by metaldetector107 View Post
    dear
    Ivconic Max quazi and all
    please i need your help
    my problem is my device works perfect it detect a 1.5cm silver coin about
    20cm with full discrimination also full reject of nail and other iron stuff
    but it detect something like disoldering and something similar no problem..
    my main problem is
    it detects my hands from few cm also it detects ferrit from AM radio antenna from same distance i have adjested GB resistor but i canot get
    lower than this distance also when moving the machin it sound like crack
    i tried many ways like reduce sens ,,i tried for RX voltage from 0 to 30 mv
    same thing so please help
    thanks in advance
    Hi,
    the hands detection is a capacitive problem: you need to shield tx and rx coils, and make so proper shield with gap like described early in this thread.

    The ferrite, instead, is different: what are dimensions of ferrite you use ?

    With small pieces and also toroids you can totally filter out it if machine is ground balanced ok. Really you don't even null totally ferrite out, cause small detection and few cms is good... you hear like cracks when passing at few distance (1-2cm is good, less is even better sometimes), but remember real soil is a little different from ferrite so even if md sounds with crack at few cms with ferrite it will not sound on normal soil; you can use TGS/TGSL on most soils but remember is general purpose md... so some soils will be hard and give false signals, that's why other models like e.g. bandidoII have manual geb pot... for fine adj on the field.

    Now, if you detect ferrite from far (say 10-15cm) it's not good: you're not ground balanced if so, so when on soil it will give alot of false signals.

    The problem could be either at coil (e.g. out from expected phase angle between tx and rx signals) or at gb trimpot setting.

    If you cannot gb using trimpot to minimize ferrite like described above you're 99% with a wrong phase angle at coil. But you need scope to fix that and match expected angle (around 200° for tesoro umax coils like tgs uses).

    You can fix with scope (the right way) or, if you have no scope, modify your coil if unglued by trial-and-error... till you can balance, or just make new one and cross fingers! Unfortunately this is a critical factor in good coil making cause TGS/TGSL require you match that angle for good operations.

    Try to use small ferrite, 5-6cm are good for rods, small toroids are good too (2-4cm diameter). You can also make a dummy thing with epoxy and powder of iron mixed to form a coin shaped thing... 3-4cm diameter for similar purposes.

    Remember: when right done you can get crack on ferrite at very few distance BUT full e.g. silver detection with loud beep in disc mode! If not you're far from good balance of your md.

    Kind regards,
    Max

    Leave a comment:


  • metaldetector107
    replied
    TGSL ground effects

    dear
    Ivconic Max quazi and all
    please i need your help
    my problem is my device works perfect it detect a 1.5cm silver coin about
    20cm with full discrimination also full reject of nail and other iron stuff
    but it detect something like disoldering and something similar no problem..
    my main problem is
    it detects my hands from few cm also it detects ferrit from AM radio antenna from same distance i have adjested GB resistor but i canot get
    lower than this distance also when moving the machin it sound like crack
    i tried many ways like reduce sens ,,i tried for RX voltage from 0 to 30 mv
    same thing so please help
    thanks in advance

    Leave a comment:


  • simonbaker
    replied
    Good fix.

    LED takes some battery power -- do you really want it?

    Cheers,

    -SB

    Leave a comment:


  • Kahuna
    replied
    LOOOOOOOL

    i solved the problem ... shame on me!!!!!

    i had a led between positiv after switch and gnd ...
    i had soldered it on next possible gnd near to led and thats in this case my geb pot.but ... i soldered it not to gnd input of pot i do it to the output pin
    i saw it some minutes ago as the led goes lighter and darker by turning geb
    i never take a look at it before ... lol

    and that little thing do so much problems

    no very quit
    detection of 1 euro coin in 20 cm
    disc work ... not fine but work
    i hope it will be fine with better coil

    after rechecking null with coil cable on it i have 23.4 mV

    it could be better

    so i'll make a better coil and i think it will be great


    thx for hlp simonbaker

    best regards

    huna

    Leave a comment:


  • simonbaker
    replied
    Originally posted by Kahuna View Post
    yes i have a digital one

    today i checked coil with LCR-meter
    TX: 5.8 mH
    RX: 6.2 mH

    should be in tollerance

    replaced both bf245c cause i thought they could be damaged by false output of lm393.
    centertap is removed

    but same problems

    can you give me some points for meassures with right datas so i can check this

    would be nice

    thx and regards
    - One thing to start with -- try reversing the two RX leads to see if that changes the double-beep to a single beep.

    - Check your supply voltages referenced to ground. V+ should be +8 volts. V- is typically -5 volts, plus or minus a volt. Check V- at the junction of D6 and C5, or any other place you see on the schematic.

    To try to trouble shoot, we need to see your waveforms. Using a dual trace scope, capture the following pictures (with a digital camera for example) using two traces (called "top" and "bottom"):

    1. TX waveform (pin 1 of J1) on top and RX waveform on bottom (pin 7 of U101a, output of RX LF353).

    2. TX waveform (pin 1 of J1) on top and ground channel driver (pin 7 of u102b) on bottom.

    3. TX waveform (pin 1 of J1) on top and all metal driver (pin 1 of J3) on bottom.

    4. TX waveform (pin 1 of J1) on top and discrim driver (pin 3 of J3) on bottom.

    It would be helpful to repeat #4 with the discrimination pot (R18 ) set in different positions, such as fully left, middle, and fully right.

    It would also be helpful to repeat #2 and #3 with the ground balance pot (R22) in three different positions also.

    I know that is a lot of pictures, but would be helpful for the experts here.

    Best regards,

    -SB

    Leave a comment:


  • Kahuna
    replied
    yes i have a digital one

    today i checked coil with LCR-meter
    TX: 5.8 mH
    RX: 6.2 mH

    should be in tollerance

    replaced both bf245c cause i thought they could be damaged by false output of lm393.
    centertap is removed

    but same problems

    can you give me some points for meassures with right datas so i can check this

    would be nice

    thx and regards

    Leave a comment:


  • simonbaker
    replied
    Originally posted by Kahuna View Post
    hi guy,



    some problems and i need some help ...

    first time i build a tgsl on selfmade pcb, at the first test i blow the lf353 and the lm 393 at rx stage by placing the 353 in opposite direction. both get very hot and i decided to replace both with new parts. and i decided to build the pcb from driscoll from page 113 with the same parts. so i removed all parts from pcb and placed them on new one. at this time i noticed that my dealer send me 2k2Ohm instead of 1MOhm. thats the prehistory

    question: what parts could be damaged in case of taking 2k2 Resistors instead of 1M and whhat parts could be damaged if i turn the lf353 at 180°?

    next ...

    i finished the tgsl on the new pcb with new 353 and 393 on rx stage and real 1M. Coil is nulled. first test...
    no tone ... i removed center tap from pcb ground and i get a tone.
    second try ...
    i switch it on and the machine starts to beep for many times. no change if i turn geb or sens pot, when i turn disc pot the beep stop (the machine is in all metall mode ?!?!?!). if i put some metall over coil i get an answer for detection and the beeps return. when i switch to disc mode i have no beeps and a good detection 1 euro at 25 cm. but the detection tone if the coin has a distance to coil from around 10cm is somtimes a double beep.
    but... disc dont work ....
    i can turn the pot in any dirction with no change. in every position i can detect aluminium, iron and 1 euro.

    so ....

    why it doesnt work with center tap on gnd
    why i have an endless beeping in all metall mode
    why id have no disc in discmode

    some ideas ? i have no more ideas ... need some help pls

    coil: wire 0.28 (dont know if it is with or without varnish)
    DD dimensions like ivconics recipe
    tx 104 , 19.7Ohm , 14,6KHz at tx stage
    rx 110 center tapped, 27,7Ohm
    nulled to 0,000V very easy

    all parts are correct connected

    THX

    Seems like you did pretty well so far. Do you have oscilloscope? Hard to troubleshoot without.

    I think center tap on RX was experiment of Ivconics, but normal design doesn't have it.

    Regards,

    -SB